Hi Dhananjay,
No problem! You can certainly use the SCB-100 with the
PXI-6514.
In fact, if you look here, you will
see that it is one of the preferred ways to connect signals to your device.
What is probably happening that is causing you not to get the data you want
though is that the SCB-100's switches are in the wrong location. If you
look in the User
Manual for the SCB-100 you will see a series of images showing you the
proper configuration. You want to set up the device like this:
Figure 4. 100 Generic
Screw Terminals Switch Configuration
with Temperature Sensor and Accessory Grounds Disconnected
This will make sure that nothing is connected to the wrong grounds, etc. And
you should be able to begin taking measurements.