Thank you for posting to the NI forums. For thermistor measurements, you will need to provide some form of current excitation to the transducer. The 9211 does not provide any form of current or voltage excitation.
Instead, I would recommend using the NI 9219, which has built-in current and voltage excitation. You could also use any of the other analog input modules, but again, you will have to make sure that they either provide excitation, or you will have to provide it externally.
I hope this helps. Please post back with any further questions.
Ed W.
Applications Engineer
National Instruments