08-05-2008 12:55 AM - last edited on 02-16-2024 11:17 AM by migration-bot
08-06-2008 11:27 AM - last edited on 02-16-2024 11:20 AM by migration-bot
Hi Ryana,
Thermocouples generate a small voltage based on their temperature. To measure a thermocouple you must be able to read this analog voltage. The PXI-6514 is a digital I/O card that has no analog input channels. That is why it can not be used to measure thermocouples.
With your current hardware you can measure a thermocouple using the PXI-4060 DMM. You can use the DMM along with your PXI-2575 mux to expand the channel count. YOu will also need a CJC sensor that the DMM can measure.
The 4060 DMM is not programmed with NI-DAQmx. You would need to use the NI-DMM driver. THe switch can be controlled with NI-DAQmx or NI-Switch. Examples for both of these drivers can be found in LabVIEW under Help >> Find Examples...
Also here are some Developer ZOne articles that explain how thermocouples work and some considerations when measuring them with a DMM.
Taking Thermocouple Temperature Measurements
Performing High-Accuracy Temperature Measurements Using a NI Digital Multimeter and Switch
08-06-2008 06:44 PM
08-07-2008 03:09 PM - last edited on 02-16-2024 11:21 AM by migration-bot
Hi Ryana,
The SCB-100 is not designed to be connected to the PXI-4060 DMM or PXI-2575 switch. To do multiple thermocouple measurements with those two products you will need a terminal block for the PXI-2575 and the wire a CJC sensor and your thermo couples to the switch's channels with the DMM connected to the switches com channel. This will let you scan through your thermocouples and CJC sensor taking a measurement of each using the DMM's single channel. If you do not have a terminal block for the PXI-2575 the NI recommended set up is 1 - LFH200 to 50-Pin DSub Switch Cable (CH-CH Twisted) cable and 4 - NI TBX-50, 50 Pin Dsub Screw Terminal Block. This setup would give you access to all 200 pins on the PXI-2575 and allow you to measure a CJC and 97 thermocouples with the 4060.
Let me know if you have any questions and take care.
08-07-2008 08:11 PM
08-08-2008 02:18 PM - last edited on 02-16-2024 11:23 AM by migration-bot
Hi Ryana,
Semiconductor sensors, thermistors, or RTDs are commonly used to measure the reference-junction temperature. Basically for a CJC sensor you just need an accurate temperate sensor at the terminal block that can be read by the DMM. Since the DMM can measure resistance I would recommend using an RTD or thermistor for your CJC sensor. Check out these two Developer Zone articles for more info on these transducer types.