Hallo,
I have a problem with an 8451 SPI interface. It is connected to my system (labview 2019) and is recognized by the Measurement and Automation explorer - but then only for use of the digital I/O ports.
When trying to use the dedicated 845x drivers with SPI support in my labview program, no device is found.
How can I make sure that the device is available for the dedicated driver and not claimed by DAQmx?
(DAQmx is required for another instrument).
Solved! Go to Solution.
05-05-2022 09:50 AM
It seems unlikely that DAQmx will claim an 845x device as DAQmx does not even support the 845x device. Could you post screenshots of the issue you're facing?
05-05-2022 11:12 AM
Drivers can be found here for the 8451:
https://www.ni.com/en-us/support/downloads/drivers/download.ni-845x-driver-software.html#428064
-AK2DM
05-06-2022 02:18 AM
Hi all, thank you all for the replies.
The 845x driver is installed, this is not the problem. However, the device is claimed by a driver called "USB-6501". I do not have a photo of the device to share today (home office with remote connection), but it is clearly labelled outside USB-8451.
Below are some screenshots: NI package manager, windows device manager, measurement and automation explorer, 845x in labview, the probem with the device not being present.
05-06-2022 10:03 AM
Are you sure it is an 8451? it definitely looks like a USB-6501
Both 8451 and 6501 use similar enclosures and connectors (maybe the same) and it could be mistaken for the other.
05-09-2022 06:53 AM
After checking again it turned out that my device really is a 6501 which was mounted in a 8451 enclosure. Quite confusing...
Thank you all for your valuable help.