Setting clock reference for USB-6001

I am trying to set up a USB-6001 to send a pre-programmed sequence of TTL pulses on software command. I am doing the programming in Python (via nidaqmx library). However, I can't seem to be able to set the rate for the sample clock. I'm able to set up a task and change the state of a pin by command, so I know everything is loaded correctly, at least.

 

 

 

>>> task_do.do_channels.add_do_chan('Dev1/port0/line0')
DOChannel(name=Dev1/port0/line0)

>>> task_do.do_channels.add_do_chan('Dev1/port0/line1')
DOChannel(name=Dev1/port0/line1)

>>> task_do.write([True, True])
1 #I can observe pins turning on

>>> task_do.write([False, False])
1 #I can observe pins turning off

>>> task_do.timing.cfg_samp_clk_timing(rate = 1000)
Traceback (most recent call last):
File "<pyshell#28>", line 1, in <module>
task_do.timing.cfg_samp_clk_timing(rate = 1000)
File "C:\Users\bglebov\Python36\lib\site-packages\nidaqmx\_task_modules\timing.py", line 3007, in cfg_samp_clk_timing
check_for_error(error_code)
File "C:\Users\bglebov\Python36\lib\site-packages\nidaqmx\errors.py", line 127, in check_for_error
raise DaqError(error_buffer.value.decode("utf-8"), error_code)
nidaqmx.errors.DaqError: Requested value is not a supported value for this property. The property value may be invalid because it conflicts with another property.
Property: DAQmx_SampTimingType
Requested Value: DAQmx_Val_SampClk
Possible Values: DAQmx_Val_OnDemand

Task Name: DO test

Status Code: -200077

 

 

From what I understand, I need to set the clock before setting up the waveform. For this, I'd like to use the internal clock of the USB-6001. The spec sheet defines the timebase frequency at 80 MHz, and max sample rate at 20 kHz. However, I get the above error for any rate value. (Just for kicks, I also tried passing a string for rate value, and it came back with a different error - so it's not a type error, at least.)

 

Does USB-6001 even have an internal clock reference, or does it have to us an external one? If so, how do I set this up?

 

Also, assuming I can set this up, would the sequence fire off whenever I run task_do.start()? I couldn't find a setting for a software trigger.

0 Kudos
Message 1 of 5
(170 Views)

The USB-6001 only supports *static* DIO, not hw-clocked and buffered DIO.  So you can't use a buffer or set a sample rate.  This remains true even if you supply an external clock.  (The 20 kHz max sample rate you found applies to AI).

 

Depending on the sequence, you *might* be able to do it with AO instead.  Max rate there is 5 kHz so your pulse timing would be limited to 2 millisec resolution.

 

 

-Kevin P

 

0 Kudos
Message 2 of 5
(153 Views)

Hm. Thanks for clearing that up. Well that's rather disappointing. And not entirely clear from either the spec sheet or the manual. Unfortunately, I was already planning to use the analog outputs for something else, so that's a no-go.

 

Do any of the other USB-600x models support clocked and buffered digital outputs? I'm looking for something in this form factor. If I end up using something much bigger (like, a cDAQ module), I might as well use a programmable function generator, which I already have.

0 Kudos
Message 3 of 5
(147 Views)

You definitely have to go for something better than USB-600x for dynamic DIO.

 

All USB-600x only support static DIO.

-Santhosh
Semiconductor Validation & Production Test
Soliton Technologies
NI CLD, CTD
LabVIEW + TestStand + TestStand Semiconductor Module (2013 - 2020)
NI STS for Mixed signal and RF
0 Kudos
Message 4 of 5
(133 Views)

The spec sheet *is* fairly clear about only supporting "static" DIO, but I recognize that while the *meaning* of static DIO is plenty clear to me who's been around NI stuff for 20+ years, it may not be so clear to anyone who isn't already accustomed to the jargon.

 

 

-Kevin P

0 Kudos
Message 5 of 5