‎01-29-2024 11:43 AM
Hello All,
I'm using PXIe 6363 (10V) to acquire data from four sensors with 10k ohms impedance each. That means I need at least 4mA current to drive the sensors (4 x 10V/10000ohms = 4mA). How much is the current rating for AI channels of PXIe 6363 and where do I find this information? Thank you.
Solved! Go to Solution.
‎01-29-2024 03:13 PM
AI stands for Analog "Input" and inputs don't supply current.
‎01-29-2024 03:18 PM
So I will have to power my sensors through external power supply? Thank you.
‎01-29-2024 04:19 PM
You can use the +5V terminals on each connector to power your sensors.
What sensor are you powering? does it have to be strictly 10V?
‎01-29-2024 04:22 PM
Thank you for the screenshot!
I have four linear potentiometers to measure displacements which need 10V (and have 10kohms impedance), so the 5V terminal might not work.