PXI 6534. Adjusting sample point

Hello.

I am using two 6534:s to send out and sample data. I synchronize them using the RTSI bus. As it is now I think I'm sampling the data at the same time I send out data. My question is if it is possible to adjust the point where the data is sampled, so that the data is sampled 20ns after data is sent.

I'm attaching the .vi I'm using.
0 Kudos
Message 1 of 2
(2,988 Views)

Hello!

 

I tried out this myself with the DAQmx driver and unfortunately I found out that what you would like to achieve is not supported on the 653x series. Since I wanted to be completely sure I verified this by calling one our our digital board specialists in the US.

 

Information about boards that are capable of this can be found here:

http://digital.ni.com/public.nsf/websearch/3C7C4BAFEFCF536586256F3C004D52C4?OpenDocument

 

http://zone.ni.com/devzone/conceptd.nsf/webmain/08f0c481d095b2ab86256ddd007724b7

 

Regards,

Jimmie A

Applications Engineer, National Instruments

 

Regards,
Jimmie Adolph
Systems Engineering Manager, National Instruments Northern European Region