I recently purchased the NI-USB-6210 to replace my old voltmeter, Keithley 197A. After installation and initial testings, I found that the baseline noise was significant increased.
For the same setting, my old Keithley 197A meter measured 5.0000+/-0.0001 V. However, the new NI-USB-6210 measured 5.00+/-0.01V. The noise is 100 times higher.
At a lower voltage, it's 0.500+/-0.002V VS 0.5+/-0.4V. The noise is too high. According to the spec of 6210, it measures 10 V in 16 bit, and this translates to 0.1 mV. So I suspect the card may be defective.
Does anyone have the same issue? Would this be defective or could it be some settings to increase the resolution?
(7,901 Views)
Several things: (this would probably be better in the Multifunction DAQ Board as it is unrelated to Digital I/O)
1. The resolution of the USB-6210 on the +/-10 V range is 20 V/(2^16) = 0.305 mV.  The resolution is calculated on the full scale range.
2. The noise level is specified on the AI Absolute Accuracy Table (page 4 of the Specifications document) as 229 uV rms. That means the peak to peak noise is probably about a mV.
3. The Keithley takes three readings per second and probably does some smoothing or averaging of the voltage during that time.
4. The USB-6210 aperature time is not specified but is probably on the order of the reciprocal of the maximum sampling rate or ~ 4us. The input bandwidth is 450 kHz.  The magnitude of the noise on a signal is proportional to the square root fo the bandwidth.  The bandwidth of the 197A is not specified for DC measurements but may be as low as a few Hz. and is quite likely less than the power line frequency. That could produce a noise reduction by almost a factor of 100.
What noise do you get with the inputs to the USB-6210 shorted to ground using very short wires?
You may want to try taking multiple samples and averaging them.  This reduces the noise, assumed to be random. Sampling at 1 kHz and averaging 333 samples would give the same reading rate as the Keithley. Also consider an external low pass filter between your signal source and the USB-6210 input.  A simple RC filter may be useful, although the details depend on your signal source and reading rates.
Lynn
(7,893 Views)
How are you wiring everything?  Using differential or RSE measurements?
What range are you using?
Lots of things can affect accuracy when you're looking for a couple of LSB's.
(7,891 Views)
I'm responding to this old post. I have experienced an unexpected noise using a DC current shunt with a USB-6210 to measure low level DC voltages on the USB-6210 0 to 200 millivolt scale. I couldn't get rig of the noise without very significant filtering but that affected the dynamic response. The noise turned out to be noise generated from within the USB-6210 itself! Obviously nothing can be done about that at least realistically. That said the solution is to use a front end DC signal conditioning amplifier and this will "quietly" amplify the signal to a much higher level such as 0 to 10 VDC and then use the 10 VDC range on the USB-6210. This vastly increases the signal to noise ratio to the point where the USB-6210 noise is so insignificant in the measurement it doesn't matter. There are at least a couple of companies out there making excellent DC instrument amplifiers that will do this nicely. Hope this helps anyone.