Our measurement setup is using usb-1208FS to read mv/V sensor outputs into Dasylab.

Provided with 15V DC excitation and 2mv/V max output from sensor --> Max output into DAQ will be 30mV. 

Analogue input module only allows a minimum of -1V to +1V of input range, this means a "theoretical" resolution of 2V/(2^12) = 0.49mV, which is pretty bad....

 

Hence my questions are:

1. Is this -1V to +1V input range specified by Dasylab or the usb-1208FS card?

2. Any suggestions/comments how can we improve this?

 

I would imagine either get a DAQ card with better resolution, maybe 16 bits instead of 12.

 

Would like to hear any comments and appreciate any inputs.

Regards

Kent

0 Kudos
Message 1 of 2
(2,538 Views)
Solution
Accepted by topic author Kentzzzz

The limitation is the device. Dasylab enables most features of supported hardware.

 

Look at the 24 bit USB-2416 for both range and resolution, or the USB-2404-UI.

Measurement Computing (MCC) has free technical support. Visit www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.