Counter/Timer
Why am I having trouble reading linear encoder at higher rates of speed?
I am trying to read linear position using a rotary encoder rolling along a surface.  My setup seems to work just fine for linear speeds of up to about 385 mm/s, however speeds beyond that cause the position value to cease changing until the velocity is reduced again.
My setup currently involves using a cdaq-9185 expansion chassis and a NI 9421 module.  The quadrature encoder is 5000 pulses per revolution and connected to a wheel of 200 mm circumference.  This results in 0.00004 meters per pulse.
I have attached an image of my code.
Given I can acquire good data at lower speeds I assume my issue is related to timing or the limits of the hardware and I am just not understanding something there.
I would like to measure substantially faster than I currently am able.  3 m/s would be great.
I would appreciate any explanation as to what I have overlooked.
(1,835 Views)
Accepted by ELDMATT
You might be up against a hardware limitation of the 9421.  The NI datasheet only specs a "delay time" of 100 usec without giving a sample rate spec.  Other searches on the site find less authoritative references to a 10 kHz sample rate limit, which is a 100 usec interval.
So I'm not sure, but that's the best guess I've got so far.  Your 385 mm/sec apparent failure point gives an encoder cycle rate of 9625 Hz, suspiciously close to the apparent 10 kHz spec limit.
-Kevin P
(1,812 Views)
With its maximum sampling rate of 102.4kHz, a NI-9361 would be able to interface your setup at speeds of up to 4m/s.
(1,796 Views)
Do your really require a resolution of 40um? If not then perhaps select an encoder with fewer ppr as an alternative to upgrading your DAQ hardware.
-AK2DM
(1,784 Views)
