Hi,

 

Got a running system with an SCXI-1104C module. I now have some measurement errors I never had before.

 

As a spare unit I have a SCXI-1104 (no "C")...  Is it direct compatible with the system ? Can I exchange

the 1104C for 1104 to eliminate multiplexor fault ?

 

//Beck777 

0 Kudos
Message 1 of 2
(3,663 Views)
Solution
Accepted by topic author Beck777

Hi Beck777,

 

You can find the manual for both the SCXI-1104 and the SCXI-1104, at the link below:

http://www.ni.com/pdf/manuals/322147b.pdf

 

If you read the common questions it explains that the only difference between the two is the cutoff frequency of the filters. It also says that  NI-DAQ version 6.7 or later is required for the SCXI-1104C whereas you only need NI-DAQ version 6.5.1 or later for the regular 1104. If you have the correct version of NI-DAQ it should work but be wary of the cut off frequency. You may end up with a lot of noise on your signal. 

 

Regards,

Tom T
Applications Engineer
National Instruments UK&Ireland