10-30-2008 07:02 AM - edited 10-30-2008 07:03 AM
To process your signals, you are going to have to become well versed in this stuff. NI does have a adaptive filter toolkit now, and maybe that might help you. Unless you can enhance your ECG detection, you are going to have problems extracting it from the noise whether you eliminate the respiration signal or not.
For what I sent you
1. Taps are the number of acquired points that are being processed each time through the loop. Essentially, each time through the loop, the oldest point is dropped out of the taps as the newest point is added.
2. Weights are the filter coeffiecients of the adaptive filtering. They are computed each time through the loop to minimize the error signal.
The error signal is the feedback that tells the filter how well its current coefficients are working. It also is your output signal. The Beta defines how much your coefficients can change from loop iteration to loop iteration. The higher it is, the more responsive your filter will be. If it is too high, the filter will become unstable and will not work. If it is too low, it will not filter as effectively as it could. The values I saved as default values are in the right ballpark, but may not be optimum.
The adaptive filter is used to help clean up your data. Until you can clean it up enough to reliably find the R peaks, it will be difficult to compute anything.
11-03-2008 04:10 AM
Dear Mr.Pursley,
can you give me some resources for the adaptive filter toolkit where I can get some information to process or enhance my signal. I looked at the examples on labview but I could not find much on signal enhancement.
11-03-2008 04:24 AM
Dear Mr. Zhijun Gu,
thanks for your mod vi, why is that I cannot run any of the other signals using your mod vi. I can only run "R10-R10_closed cover_Bhavin1_1.lvm" signal here and that too its not continous it just scans for a few seconds and then stops, how can I make this to run continously.
thanks for your time
11-03-2008 09:34 AM - edited 11-03-2008 09:38 AM
thanks for the link, due to short time constraints it is not possible for me to work on the hardware signal enhancement, and my task is to extract the best signal conditioning from this hardware, so I am really bound by time and trying my best to get HRV out of this signal.
I have downloaded the trial version of adaptive filter toolkit..its for 30 days only..although my signal is full of noise but in some of the previous recorded signal that I have sent, I can see ECG, I know its not very clear, but I have to work with them only, as I dont have any option at the moment.
From the attached screenshot you can see some ECG signal but my I am not sure weather I am getting the correct measurements here..your previous data mod vi works absolutely perfect with the conventional ECG, but with my signal it looks as in screenshot. I have a doubt with the locations of Heart rate, when I change the location to 1,2,3,4,...and so on it shows different heart rates every time..so I am confused by this, can you explain me what exactly is happening here..I have to display a mean heart rate, and mean value of R-R interval at the end...how can I do this.
thank you for your time,
abi
11-03-2008 10:14 AM
It looks like you are getting the right signal, but you may have to adjust some parameters to optimize it. You might try to increase the width parameter of your Peak Detector.vi to avoid detecting the double peaks. You should also play with the filters and wavelet stuff to see if a different wavelet or cutoff frequency works more effectively.
Instead of getting a heart rate measurement from each R-R distance, get a single number for each 5 second set of data. In addition, you know that heart rates should be between let's say 50 and 300 bpm. Any number outside the range should be thrown out before averaging the distances from the remaining peaks.
You might even try to average across a 10 second span to get more peaks for an average. Create a 10 second buffer of data and each time through the loop drop off the oldest 5 seconds worth of data and add the new 5 seconds worth of data.
11-14-2008 07:36 AM
Here is some info on an article that just came out that you might look into.
Tonometric Arterial Pulse Sensor With Noise Cancellation
Ciaccio, E.J.; Drzewiecki, G.M.;
Biomedical Engineering, IEEE Transactions on
Volume 55,
Issue 10,
Oct. 2008
Page(s):2388
-
2396
Abstract:
Arterial tonometry provides for the continuous and noninvasive
recording of the arterial pressure waveform. However, tonometers are
affected by motion artifact that degrades the signal. An arterial
tonometer was constructed using two piezoelectric transducers centered
within a solid base. In two subjects, one transducer was positioned
over the radial pulse (p) and the other was positioned on the wrist not overlying the pulse (n).
The presence of induced motion artifact and any noise was removed after
signal digitization by noise cancellation. Besides fixed weighting, two
adaptive algorithms were used for cancellation-LMS and differential
steepest descent (DSD). Criteria were developed for comparison of the
adaptive techniques. The best fixed weighting for noise cancellation
was w = 0.6. For fixed-weighting, LMS, and DSD, the mean
peak-to-peak errors were 1.22 ± 0.54, 1.18 ± 0.30, and 1.16 ± 0.23 V, respectively, and the mean point-to-point errors were
15.86 ± 3.15, 11.40 ± 1.96, and 10.13 ± 1.25 V,
respectively. Noise cancellation using a common-mode reference input
substantially reduces motion artifact and other noise from the acquired
tonometric arterial pulse signal. Adaptive weighting provides better
cancellation than fixed weighting, likely because the mechanical gain
at the transducer-skin interface is time-varying.
11-14-2008 07:39 AM
11-17-2008 07:18 AM - edited 11-17-2008 07:22 AM
Dear Mr. Pursley,
I have modified a vi based on your previous adaptive filter concepts, I really dont know if this makes sense here. Kindly have a look at my attached files.
I have 3 data samples here (Real ecg) that is conventional ecg and the other two are samples from my real time measurements with DAQ device.
In the second data sample R10-R10_closed cover_Bhavin1_2, the output after wavelet denoising looks good, only thing is I am not able to get the
heart rates and the intervals here. Kindly have a look at it.
thanks for your time,
Regards,
abi
12-05-2008 10:43 AM