From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

DASYLab

cancel
Showing results for 
Search instead for 
Did you mean: 

Data slow to update

I am trying to shorten the time that it takes for my current data to display on my digital meter. Currently the data is updated every 3-4 seconds. I would like to see it in real time. I have experimented with sample rate and different block sizes without any success. My system includes a PersonalDaq 56 coupled with Dasylab software. I am only utilizing one channel to capture voltage, the data is sent to a scaling module and then to a digital meter in addition, I have another digital meter that is connected directly to the output of the data logger. Any assistance will be greatly appreciated.

 

Thanks,

Wayne

0 Kudos
Message 1 of 5
(6,161 Views)

The only way to improve the sample rate is to disable the calibration option and then to change each input channel to have a shorter sample duration. 

 

When using the PersonalDAQ56, the sampling rate is determined by several factors. In the Measurement setup for the device, you can select the auto-calibration mode and the overrange protection. The calibration mode takes time at the start of measurement before the first sample is collected and periodically during the measurement. 

 ScreenShot369.jpg

Then... the actual sample rate maximum depends on the measurement duration (per channel, set in the Analog input module) and the sample rate (set in the Measurement setup for the device).

 

ScreenShot370.jpg

Measurement duration (per channel) – the amount of time used for sampling a channel’s input signal.
You can independently set the measurement duration for each channel. The measurement durations for
Personal Daq’s analog channels range from very slow (610 milliseconds for one sample) to very fast
(12.5 milliseconds for one sample).


Sample rate – Samples per second. The sample rate is the number of samples that take place per second.
With the very slow measurement duration of 610 milliseconds, there will only be 1.6 samples per second.
With the very fast measurement duration of 12.5 milliseconds, there will be 80 samples per second.

 

The manual for this device has a nice table, showing the impact of the various settings. 

 

ScreenShot371.jpg

Measurement Computing (MCC) has free technical support. Visit www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.
Message 2 of 5
(6,160 Views)

Hello CJ,

 

Been along time since we last talked, Thanks agin for all your help getting our software program up and running. Our new adventure is reading smoke from our smoke generator. I am trying to get the readout to update faster and speed up our test runs. It makes it a lot easier to run the equipment if you can see the changes in real time without the 3-4 second lag time. I have made a new worksheet with only the PersonalDAQ module and a digital meter module to  try to figure this out. Currently I have the sample rate at 40 msec, I've turned off the auto calibration and the overrange protection but nothing seems to help. I have used the PersonalDAQ software to see if my PC is the issue, it seems to work just fine but I could not find any scaling capabilities so that is not an option for us. If I can get the data to update once per sec I would be very happy. What would you suggest I try next. Thanks

0 Kudos
Message 3 of 5
(6,157 Views)

Did you go to the Measurement Menu and increase the sample rate for the PersonalDAQ? 

 

Otherwise, you'll need to post the worksheet, and let me try it here. 

Measurement Computing (MCC) has free technical support. Visit www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.
0 Kudos
Message 4 of 5
(6,155 Views)

Problem solved, had to go into PDAQ acquisition settings and manually type in the sample rate. Thanks again for you assistance

0 Kudos
Message 5 of 5
(6,153 Views)