Hello,

 

I work in a secured area and we recently purchased 2ea of the VXI-PCI8026.  Can someone direct me to a Letter of Volatility for the EEPROM on the VXI-MXI-2 and PCI-MXI-2 module/board?  Also, what exactly is the size of the EEPROM, I couldn't find this mentioned anywhere in the manual.

 

 

Thanks,

- Darin

0 Kudos
Message 1 of 5
(8,625 Views)

Darin,

 

Thank you for using the Forums.  I believe the document that you are looking for can be found here.  

 

I hope that helps.

 

Brian P.

 

Applications Engineer
National Instruments
Message 2 of 5
(8,615 Views)

Hello Brian,

 

Thank you for the quick reply.  Does that Letter of Volatility apply to the VXI-PCI8026 also, even though it states it's for the VXI-PCI8015?  If that's the case, I'm assuming it applies to any VXI-PCI80xx and that the only difference between the VXI-PCI8015 and VXI-PCI8026 is the software package that ships with the module and board?  Is that correct?

 

 

Thanks,

- Darin

0 Kudos
Message 3 of 5
(8,612 Views)
Solution
Accepted by topic author DarinM

Darin,

 

You are correct, I have checked the part numbers for both boards and the actual letters of volatility and they are the same.  I am sorry for the confusion.  I hope that cleared it up.

 

Regards,

 

Brian P.

Applications Engineer
National Instruments
Message 4 of 5
(8,608 Views)

Excellent!  Thank you very much!

 

 