From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

PCI-6221: Bad voltage reading when reading all channels at once

Solved!
Go to solution

Hi!

 

We have a PCI-6221 acquisition card which has a very strange behavior. Channels 1 & 2 are connected to an 8V source. When performing an acquisition on those channels only (RSE mode), everything is perfect. But if we perform the acquisition on all the channels at once, there is an offset in the measured voltage.

 

I include a screenshot of the problem with NI MAX. On one file, only channels 1 & 2 are acquired and we measure 8V. On the other file, all channels are acquired at once and we can see than channel 1 (red line) is around 7.75V, while channel 2 (green line) is correctly at 8V.

 

Does anybody know why this happens and how to fix this?

 

Thank you

Download All
0 Kudos
Message 1 of 3
(2,356 Views)
Solution
Accepted by topic author brunoobsomer

Ghosting (settle time error) of multiplexed DAQs!

See spec diagram called 'settle time error' , the scale is in ppm but log!

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 3
(2,338 Views)

Thanks for the hint!

0 Kudos
Message 3 of 3
(2,310 Views)