Good morning,

 

 

I have a project where I am using a PXIe-6341 that is reading a 0-10VDC signal from an instrument. The end user only wants to take the ADC "counts" for the values, we are not scaling any of the values. 

 

I have the virtual channel set for Max 10, Min 0. According to everything I have read, this should put the channel into a uni-polar mode and a gain of 1. 

 

When we collect data, we are seeing the maximum value of 32767 for a saturated value of 10V that we expect for a specific scenario. 

We were expecting to see values from 0-65535.

I do have the channel read set for RAW 1D U16, so I should be getting unsigned numbers from 0-65535. 

 

Am I missing something? We dont have any values going negative in this case but it appears that the board is only acquiring data in 15 bits and not 16.

 

Thanks,

 

Steven

0 Kudos
Message 1 of 3
(2,543 Views)
Solution
Accepted by topic author StevenHowell_XOM

A quick look at the summary spec sheet seems to only list bipolar voltage ranges.   Though a little surprising, I think the most likely explanation is that you don't get the option to spread 16 bits across the 0 to +10 volt range, you're stuck spreading them across the -10 to +10 range.  Hence only 15 bits for the positive side.

 

 

-Kevin P

CAUTION! New LabVIEW adopters -- it's too late for me, but you *can* save yourself. The new subscription policy for LabVIEW puts NI's hand in your wallet for the rest of your working life. Are you sure you're *that* dedicated to LabVIEW? (Summary of my reasons in this post, part of a voluminous thread of mostly complaints starting here).
Message 2 of 3
(2,515 Views)

Kevin,

 

Thank you for taking the time to answer my question. I looked all through the NI website and I was not able to find the answer to this simple question.

 

 