Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

Max AO Waveform with USB6363

Solved!
Go to solution

I am looking at the spec sheet for the USB 6363 and I am unsure if I can do what I want to which is A 10kHz waveform with at least 20 points per wave, and it would run for 0.5 seconds (going between 0 and 3V).  The relevant specs are below:

Maximum update rate (simultaneous) 1 channel 2.86 MS/s
Timing accuracy 50 ppm of sample rate 
iming resolution 10 ns

In the test panel, if I try to do a rate of 5000000, that says a wave of 5kHz is generated, and an error pops up saying I need to reduce.

0 Kudos
Message 1 of 2
(2,328 Views)
Solution
Accepted by topic author jmaslek

Hello, jmaslek! 

From your parameters, I calculated that you would need a sampling rate of 200 kHz.

CalcSamplingRate.PNG

That’s within spec for the USB-6363 for single channel and up to five multi-plexed channels. Where did you get the number 5,000,000 Hz?