From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

Instrument Control (GPIB, Serial, VISA, IVI)

cancel
Showing results for 
Search instead for 
Did you mean: 

GPIB-USB and Matlab

What is an 89090A. A search of Agilent's site did not return any results.

 

Can you attach the programming manual or something that explains the command you are sending? Typically, a query is terminated with a ? so the read timeout would be expected, though the fact that it does not support *IDN? could mean that the instrument has no standard programming interface at all. You obviously can run only queries that are listed in the manual.

0 Kudos
Message 21 of 51
(3,248 Views)
0 Kudos
Message 22 of 51
(3,243 Views)

Okay, I searched test and measurement.

 

From your screen shot, it does not look like you are sending the correct terminator. By default, only EOI is sent.

0 Kudos
Message 23 of 51
(3,241 Views)

So what should I do? What am I supposed to send?

0 Kudos
Message 24 of 51
(3,238 Views)

Append a LF or CR/LF as described in the manual.

0 Kudos
Message 25 of 51
(3,227 Views)

Sorry I am not familiar with the abbreviations ... what does that mean??? can you point me to the page of the manual?

0 Kudos
Message 26 of 51
(3,223 Views)

ok you mean this? What am I supposed to do with it?

 

I typed in "REM ON" and still gives me the error

 

Parameter 
[ on | off ] 
Reply 
[ on | off ] <CR> <LF and EOI>
Description 
This instruction is used to switch remote mode on and off. With REM ON the 
temperature control unit is set in a remote mode. In this remote state the 
keyboard is disabled for all inputs (only the UP , DOWN , and CLEAR  keys are 
accepted). The ENTER  key is only accepted within the remote function context, 
that is to change the display for system and main functions and to reset 
remote to off. However, the function REMOTE ON, which is selected, can be used 
to reset to REMOTE OFF instead of using the REM OFF command, (see also 
“Remote Mode” on page 88). 
Reply Description 
The REM instruction without parameters returns a string which represents the 
current state of the temperature control unit:  
on = remote mode 
off = normal mode  
See “CSM – Agilent ChemStation Mode” on page 98 for interpreting the reply 
when using the CSM command. 
0 Kudos
Message 27 of 51
(3,210 Views)
In MAX, use \r and \n for carriage return and line feed.
0 Kudos
Message 28 of 51
(3,206 Views)

Where am I supposed to edit the codes? I still don't understand... 

 

I'm still referring to this error btw:

 

Screen Shot 2012-02-14 at 2.44.50 PM.png

0 Kudos
Message 29 of 51
(3,202 Views)

I use VISA but I believe the 488 tool works the same way - simply type in the control characters after the command string.

 

You also have the EOS configuration there and on the GPIB properties page.

0 Kudos
Message 30 of 51
(3,199 Views)