Calibration Due Date

Solved!
Go to solution

 

In MAX, there is a field for "Calibration Due Date", but "Get Calibration Information.vi" provided in the VirtualBench library doesn't have an output for it.  Is this value only available by adding the calibration date and interval together?

 

Calibration_In_Max.PNG

0 Kudos
Message 1 of 2
(3,994 Views)
Solution
Accepted by KipK12

From the VirtualBench API, yes, you must construct the recommended next calibration date from the last calibration date and the calibration interval.

 

You can also get the recommended next calibration date by using the System Configuration API. There is an example that demonstrates this in the Example Finder: Hardware Input and Output -> System Configuration -> Calibration Audit.vi

——
Brandon Streiff
ni.com/compactdaq · ni.com/daq