The type of error that came up in the NIswitch_32.dll

Using the "NIswitch initialize with topology" and the 1-wire mode i get the following error:

"niSwitch Initialize With Topology.viNo legacy device supports the specified topology. Use a different topology, or configure the device under DAQmx and use the DAQmx resource descriptor.

Status Code: -1074118644 "
0 Kudos
Message 1 of 2
(6,434 Views)
If you use the SCXI-1127 as a traditional DAQ device, you can only select certain topologies. We have this note in the Help file, under Device Overview>>NI SCXI-1127:

If you configure the SCXI-1127/1128 as a Traditional NI-DAQ Device, you can only select the 2-wire 32x1 multiplexer, 2-wire 4x8 matrix, and independent topologies. To use the SCXI-1127/1128 as 1-wire, 2-wire, and 4-wire multiplexer, select the 2-wire 32x1 multiplexer topology and configure all channels appropriately in MAX.
0 Kudos
Message 2 of 2