Switch Hardware and Software
Is it possible to simultaneously connect several PXI-2567 channels with one software trigger?
I hooked up a 4 channel scope to the 4 resistive loads and measured the time delay between each turn on.
Message 11 of 16
(2,399 Views)
Here's the scan list command I setup... I actually used 8 channels at the same time in my final solution, not 4.
";ch24->com24&ch25->com25&ch26->com26&ch27->com27&ch28->com28&ch29->com29&ch30->com30&ch31->com31;~ch24->com24&~ch25->com25&~ch26->com26&~ch27->com27&~ch28->com28&~ch29->com29&~ch30->com30&~ch31->com31;"
Then sent a software trigger command to enable this at the card level.
Message 12 of 16
(2,395 Views)
So it looks like you turned on 8 relays and then after they have all turned on, you turned them all off.
I wonder if you would have entered your command like this:
";ch24->com24&&ch25->com25&&ch26->com26&&....;~ch24->com24&&~ch25->com25&&~ch26->com26&&....
I bet it would have handled each relay in order. I wonder how much time it would take to wait for the debounce though?
Message 13 of 16
(2,391 Views)
Now I may be understanding it wrong, but looking at the spec sheet of the NI PXI-2567 under Dynamic Characteristics, it says that single channel operation time (typical at 25 deg C) is 60us.
And i don't need to open AND close the channels all within 5ms, I just need to open them in order within 5ms.
Message 14 of 16
(2,390 Views)
Sorry, I'm remembering this all in pieces.  That scan list was setup for two software triggers.  We send the first software trigger to turn all 8 channels on, wait 100ms, then send another software trigger command to turn all 8 channels off.
That's why it looks like I turn them all on then all off.
I didn't have to worry about debounce since I was using the card more like a solid state switch instead of a relay coil driver.  Debounce spec would be a parameter of the relay you're driving, not the FET driver that's energizing the coil.
Message 15 of 16
(2,386 Views)
its okay. So really, you just turned on 8 different channels "at the same time" within 10us, waited, and then turned them all off, correct?
Message 16 of 16
(2,383 Views)
