08-22-2014 08:19 AM
I'm using PXI-4071 and a PXIe-2575 in handshaking mode with a scan list.
The DMM is measuring current and the MUX is in 1x196 mode.
When running the scan list, I rarely (randomly) get a very small measurement in the nA to uA (when it should be 4-20mA). I'll run the scan list again and that channel will be correct, so it doesn't seem to correspond to a channel.
I've tried increasing the settling time of both the DMM and the MUX, but neither have had any effect.
Any help would be great.
Caleb Swieson
Solved! Go to Solution.
08-22-2014 10:02 AM
One more point. I am reading the measurements from the DMM while it is still running through the scan list (so I can process the data as it's being measured). At the moment this is where I feel like the issue might be, but have no idea if and why that would be.
08-22-2014 10:51 AM
Cswieson,
Is the relay board relatively old? If a board is close to end of life, there is a chance that the relay will sporatically not close and/or open. If the relay is stuck open, the DMM might measure very low current because no current is flowing (and only noise is coming through).
08-22-2014 11:03 AM
Hi Frank,
No, the board is only a year old. I checked the relay counts and the most counts is still <10,000. Most are less than 1000.
Caleb
08-28-2014 11:22 AM
After further testing, we have determined it was actually the devices being measured that were outputting the very small currents. So it wasn't the hardwares' fault. Whoops
08-28-2014 02:07 PM
Thanks for following up, cswieson! I'm glad you were able to find the root of the issue.