I am using the two SCXI-1193 rf switch configured as a Quad 8x1 Mux and I'm using NISE v2.0.1. I have created a hardwire that connects from ch0 of rfSwitch_1 to com0 of rfSwitch_2. So, for example, when I go to create a route from rfSwitch_1 com0 to ch3 of rfSwitch_2, I get the following error:
"Channel or repeated capability name specified is not valid for the instrument."
Why am I getting this error message. It seems to me that since there is a hardwire created between the two rfSwitches there shouldn't be any problems connecting a path from rfSwitch1 to rfSwitch2 between the two quad mux's.
I am using the two SCXI-1193 rf switch configured as a Quad 8x1 Mux and I'm using NISE v2.0.1. I have created a hardwire that connects from ch0 of rfSwitch_1 to com0 of rfSwitch_2. So, for example, when I go to create a route from rfSwitch_1 com0 to ch3 of rfSwitch_2, I get the following error:
"Channel or repeated capability name specified is not valid for the instrument."
Why am I getting this error message. It seems to me that since there is a hardwire created between the two rfSwitches there shouldn't be any problems connecting a path from rfSwitch1 to rfSwitch2 between the two quad mux's.
02-17-2006 04:08 PM
02-17-2006 05:23 PM
02-17-2006 05:37 PM - edited 02-17-2006 05:37 PM
The correct device needs to appear in the Hardware tab inthe Hardware asset window.
When you configure the 1193 in quad 8x1 each channel can only be routed to its corresponding com. Please check the NI Switches help file under devices>>SCXI-1193>>Quad 8x1 multiplexer topology.
The error: "Channel or repeated capability name specified is not valid for the instrument" means that you are trying to use channels that do not exist in the switch device you are trying to use. The IVI configuration has to be incorrect in order to get this error.
Message Edited by Guy Ferraro on 02-17-2006 05:38 PM
02-17-2006 05:45 PM
02-20-2006 11:15 AM
I am convinced that the IVI configuration is corrupted which prevent you from using the SCXI-1193. If you do not have any hardware you would have had to simulate both SCXI-1193. The fact that you are seing all the channels in NICE virtual device but are getting error when trying to use them tells me that somthing changed in the IV configuration and that you are not using the initial topology used to read all the available channels for the 1193.
If you do not have any hardware connected to your computer then you need to follow the procedure below to ensure that the SCXI-1193 will work as expected.
First create (2) SCXI-1193 simulated in MAX:
1) Right-click Devices and Interfaces and select Create New.
2) A dialog box prompts you to select a device to add. Select NI-DAQmx Simulated Device and click Finish.
3) In the Choose Device dialog box, select Switch>>NI PCI-4021 (the switch controller is not important but you need one to be able to si mulate a chassis).
4) Create a SCXI-1000 Simulated device using the step 1) to 3)
5) Select the device created in 3) as the communicator
6) When the chassi dialog appear, select SCXI-1193 in slot 3 and 4
7) In the chassis Tab select 3: SCXI-1193 as the Communicating SCXI Module Slot. By default the SCXI-1193 are configured as Quad 8x1
Now you can create a NISE virtual device just like if you had (2) SCXI-1193 connected to your computer. Note that logical names and driver sessions have been created for both SCXI-1193 (need to close MAX and reopen) with correct entries. If you follow this procedure you will not encounter any more issue working with simulated hardware.
Guy Ferraro
Switch Product Support Engineer
NI - USA