05-27-2014 06:50 PM
Hello guys,
Have a problem with signal express. I am just building a simple test.
As Figure 1 shows, target singal is normal sine wave with 1V offest.
Figure 1
Figure 2 shows that the offset effect quickly dishminished on the generated signal (detected).
Figure 2
I am very confused on this phenomenon. Anyone can help give an explanation?
If I want to a normal sine wave with steady offset, what should I do?
Thank you
Kevin
Solved! Go to Solution.
05-28-2014 12:13 PM
Hey Kevin!
The reason that's occuring is because you are using AC coupling on your input, so the DC component is being filtered out. The PXI-4498 is an AC coupling only device so unfortunately you won't be able to read that 1V offset on the input.
http://digital.ni.com/public.nsf/allkb/0F81C33977056A7D862566F1005CAA71
The following user guide is where it states the 4498 can only perform AC coupling:
http://www.ni.com/pdf/manuals/372125f.pdf
Hope this helps clear things up!
05-29-2014 12:19 PM
Thanks Rob,
It helps.
Kevin.