I'm using a NI USB-6210 with SignalExpress to take temperature readings from a type K thermocouple. The readings are all over the place. If I add more channels to aquire from, the readings get within a 2 degree C range, which is close to acceptable. Still seems like a large swing for a thermocouple device that is sitting in a room close to room termperature. Is there anything I'm doing wrong? I don't see anyother way I can hook things up.
Thanks
Chip
02-18-2009 07:39 AM
Since the USB-6210 is not specifically designed to read thermocouples, there are a plethora of things that could be going wrong. Without your project and a connection diagram, we can only guess. Here are a few things to consider:
02-18-2009 11:01 AM
I hope this can help you help me.
Thanks
Chip
02-18-2009 11:02 AM
02-20-2009 08:07 AM
Looking at a specification chart for the type K thermocouple, it appears that the response is about 40 microvolts per degree C (note that iron/copper-nickel is a type J, not a type K, but responses are similar). Looking at the specs for the NI USB-6210, if set to the lowest gain, it has a resolution of 4.8 microvolts and random noise of 12 microvolts. It should work OK for you. Your connection diagram is correct, but you list a type J, not type K thermocouple. Change the thermocouple type to type J (assuming you have an iron/copper-nickel). Also verify that the thermocouple itself is well bonded at the tip. Verify that your connections to the device are solid.
If this doesn't work, let us know.
02-20-2009 09:42 AM
I checked the connections and they are solid. I also checked readings with type J, K settings and different thermocouples. The thermocouple gets the proper readings on another device. Still getting the same readings.
Thanks
Chip
03-02-2009 07:00 AM
Hi Chip,
I noticed that you have not received any response on your last post so I wanted to provide some more help if you still are having trouble. One thing that I noticed was that you are using a differential connection with your thermocouple (AI0-AI8, AI1-AI9). It is recommended that you use a single ended ground referenced configuration (RSE) when you are measuring a floating signal such as a thermocouple. This could be what is causing the erratic behavior on your measurement. I attached a jpg of the different wiring configurations.
Also, what is the other device that you are using that gets a proper reading with the thermocouple? What are the settings and how do you have it connected?
Field Wiring and Noise Considerations for Analog Signals
http://zone.ni.com/devzone/cda/tut/p/id/3344
03-03-2009 10:51 AM
The other device is a hydra bucket. The researcher has decided to go ahead and purchase a hyrda bucket so I no longer need help on this issue. The USB-6210 has worked great for the temperature and humidity readings I'm taking via voltage. Thanks for everyone's help on this issue.