06-29-2015 11:37 AM
I am currently attempting to use SignalExpress 2014 to acquire data from a Yokogawa WT3000 power analyzer. I have found the IVI drivers for this instrument here, but it only includes drivers for serial and GPIB. I am trying to communicate over Ethernet.
I was able to add the instrument to MAX (Network Devices -> Create New VISA TCP/IP Resource -> Manual Entry of Raw Socket with IP and port 10001), and the device can be seen. However, when I try to acquire data in SignalExpress (Acquire Signals -> IVI Acquire -> IVI Scope Acquire -> Create new sessions with TCP device from MAX and niScope driver) I get the following error:
"Configuration Error: Insufficient location information or the device or resource is not present in the system."
Any thoughts? I am fairly new to SignalExpress and have never used IVI drivers before, so it is possible I am missing something. Or perhaps it is just the fact that only GPIB and serial are supported? Is there any way for me to convert the driver so that Ethernet is supported?
Solved! Go to Solution.
06-30-2015 01:46 AM
Hi,
The WT3000 can operate just fine with the Ethernet interface.
Only it's a pity that the drivers at the NI site are not up-to-date.
Try the drivers from the attached ZIP. This one support Ethernet.
Kees
06-30-2015 10:04 PM
Thanks KC, I appreciate the reply. I actually found that driver a couple days ago, and was able to use it to acquire data via LabVIEW. But, I don't see a way to use this driver with SignalExpress. I'm assuming your comment means there is a way to use this driver with SignalExpress?
07-01-2015 01:48 AM
The link in your post points to a LabVIEW driver. I posted the driver for LabVIEW that is OK for Ethernet assuming that this will work for you.
I don't know anything about SignalExpress but I think this is only for NI hardware and IVI compatible devices. I don't think the WT3000 driver is IVI compatible.
Kees
07-16-2015 05:27 PM
For anyone else who is struggling with this issue, I found a temporary solution. Because SignalExpress only supports IVI drivers (and not plug-and-play drivers), I wrote a LabVIEW VI that uses the Yokogowa supplied plug-and-play driver to query data from the WT3000 and then publishes this data to shared variables. SignalExpress can then read these shared variables in lieu of reading data from the instrument directly.