I have a CompactDAQ with a 9233 IEPE module as well as a 9201 analog input module. I would like to measure a weighted triaxial vibration (running rms or rms) and one voltage over time, then save this to a file. I only need to get a sample once per second.
I have a couple of issues:
1) In order to measure vibration, I need to set the sample rate to something such as 25KHZ. Signal Express forces me to keep that sample rate for the analog input module also. Is this ok?
2) I can't seem to find a function that allows me to just take a reading of voltage (9201) and vibration (9233) once per second then save it to a file. The file save dialog does not seem to like mixing the signals.
3) Running at 25KHZ, whenever I try to save the voltage, it literally saves 25000 readings every second. I just want one value.
Any ideas? Again I just want the running vibration (three axes) and a single voltage saved to a file once every second.
If I need to, I can move to Labview but I was hoping to do this in Signal Express.
Jeff
12-14-2009 08:44 AM
Ok sorry I figured it out 🙂
I just used amplitudes and levels to get a DC value and I calculate the running rms vibration once per second, then I append to the file so each line is at one second intervals and consists of X,Y,Z vibration rms as well as a DC voltage.