I would check that you are following the steps outlined in the this tutorial- http://www.ni.com/tutorial/3377/en/. Be sure that you have associated the correct Driver Session associated with the Logical Name for your 5402. Also be sure to save the IVI configuration after making changes.
I also noticed that you have NI-FGEN Configuration and Development Support 14.0, but Runtime 15.0. This could be causing issues with using the 5402 through IVI with LabWindows/CVI. I would reinstall NI-FGEN 15.0 and ensure that the Configuration and Development Support are updated to 15.0 so that all three parts of the driver are 15.0.
Hope that helps,
-Eric H
01-08-2016 01:13 PM
It turns out that the line of code that initiates generation is missing from the IviFgen example code. To resolve this, please add the following line of code after the IviFgen_ConfigureStandardWaveform function:
/* Initiate output from the Function Generator */
checkErr(IviFgen_InitiateGeneration (instrumentHandle));
01-08-2016 01:36 PM
Thank you Mena
01-08-2016 01:48 PM
Nice find, Mena! herbfes, you should set Mena's post as the solution.