Dear mhtl,
I believe that your problem is grounding the signal twice. When you attach the thermocouple to the grounded machine frame, you are grounding one side of the thermocouple through the frame. You are correct in making sure that your thermocouple is grounded, but if it is grounded more than once, a
ground loop can introduce noise. Even a few mV of noise can cause tens or hundreds of degrees difference, as you are seeing.
Make sure that you are not shorting both leads of the thermocouple when you attach it to the frame. If you are attaching the thermocouple such that one of the leads electrically connects to ground through the frame, consider the
thermocouple to be grounded, and do not connect the SCXI-1100 CH� to ground in the terminal block (or use the internal W1 jumper).
If this does not solve your problem, please let me know the results from trying without the ground connection in the terminal block and we can continue troubleshooting from there.
Thank you,
Kyle Bryson
National Instruments