Suresh,
Well, you could not configure a different scan rate if channels from both modules are in your scan list and/or DAQ task.
What I mean is that you would have to use two instances of AI Config, AI Start, AI Read, and AI Clear. You would be configuring one task for the SCXI-1520 and one task for the SCXI-1102 so that each could be configured for a different sampling rate.
Regards,
Justin Britten