Thermocouple readout with scxi 1102 and labview 7.0

I am trying to read K-type thermocouple with scxi-1102 and 1303 therminal block. My problem is that the read out temperature is 2 degree lower then the onr read out with a calibrated thermocuple pt100 with another DAQ system. I think is a problem in compensation. At the moment i am using an internal compensation. How can i check that the compensation is correct? Do you have any other hint?
0 Kudos
Message 1 of 2
(2,990 Views)
There is a VI if I remembered correctly that checks the the reading of the built in compensation. I believe it is called ACQ CJC Temperature.vi found in examples. It might be what you are looking for.

As far as the reading difference goes, there is an error from thermocouple to computer screen that you might want to keep in mind. The % error of the thermocouple, the % of reading error of PCI card and % error of SCXI module. All this error can be corrected with the VI though to compensate.

Staqnok
0 Kudos