Ni 9237 problem with labview


i attached a print screen for the problem the first capture show how the DAQ read from the load cell using measurement and explorer, the second capture show how there is no read or response from the DAQ reading from the load cell using labview .
i am using NI 9237 and NI cDAQ 9174. every thing was working very good for days this happened suddenly without I change any thing .
Download All
0 Kudos
Message 1 of 3
(4,435 Views)

Hi Ahmed_WT,

 

Do you see the same behavior when you use the Bridge-Continuous Input.vi example in the NI Example finder?

 

If you are not familiar with the NI Example Finder you can access it by going to the Help tab in LabVIEW and selecting Find Examples.

The specific example that I'm referring to can be found in Hardware Input and Output>>DAQmx>>Analog Input>>Bridge-Continous Input.vi.

 

Regards

Carli S.
Troubleshooting & Maintenance Digital Business Manager - NI
0 Kudos
Message 2 of 3
(4,386 Views)

Dear Thank you for your help. the problem was because scaling.

0 Kudos
Message 3 of 3