Hello nos,
I just saw this post and thought I'd see if I can still help.
The only limitation that I believe might really affect the 9237 is the 150mW maximum power as stated on page 16 of the manual.
With this in mind, I would expect that it all depends on what excitation voltage you use.
If you use the formula P=V2/R with V as the excitation voltage and R as your device's resistance, then you should be able to check if you exceed this power limitation.
I hope this helps,
Let me know your thoughts,
Kind Regards,
Michael S.
Applications Engineer
NI UK & Ireland