04-25-2013 10:34 AM
Hello,
I have project of control DC motor in LV with CRIo in SCAN MODE and I want to scan voltage and current by analog input modul NI 9215. But I have wrong values if check it by volt-ampermeter. I suppose that is some problem with sampling of A/D converter. Because voltage and current changes very quickly and it sample very diffrent values. So the result is not correct. I just need measure effective values and make a graphes.
Thanks for advices.
Ross.
04-26-2013 07:37 AM
Hi zabloros,
could you please describe a little more in detail what and how you are measuring with the 9215 and especially with the 'volt-ampmeter'? Is it a multimeter with RMS display option?
What is your sampling frequency when using 9215? In what range and how does the sampled data change?
I suspect that you see the AC content of a signal when sampling with 9215, which is fitlered out by the multimeter when measurign with that one.
You could also try to use the Basic Averaged DC - RMS vi from the Signal Processing->Waveform Measurements palette and see if it helps.
Regards,
04-26-2013 08:38 AM
Thank you for reply. RIght now I´m working on it in lab. Sorry, my description was a bit confused. Yes, I checked voltage and current by multimetr. I have voltage and current probe connected to DC motor and NI 9215. I want to measure these two values. The voltage is twice higer then on multimetr and sometimes jumps to zero. The Basic Averaged DC - RMS doesn´t help. I just tried it. Where I can change sampling frequency? I think that I need lower frequency because it samples every disturbances.
Thank you.
04-26-2013 09:39 AM - edited 04-26-2013 09:40 AM
Settings of time engine I have 100 ms. I´m adding the grapgh of the voltage. It should be constant 15 V, which I measure by multimetr.
Thank you.
04-29-2013 04:04 AM - edited 04-29-2013 04:05 AM
Well, the attached screenshot looks good to me, this is how your signal actually looks like. With the multimeter you obtain different values, most probably because of the low-pass filtering (and perhaps rectifying too) characteristic of the device.
Sampling with lower speed will not help to obtain better results. I think, you should rather do some processing on the digitized signal (depending on your application of course).
Regards,
06-24-2015 04:03 AM
HI,
I am measuring 2 voltage signals using 9215. The cRIO is working in scan mode. I have observed that if one signal affects other signal. The rise in signal 1 bring a drop in signal 2.There are no grounding issues with the sustem.
Can you please tell me how can I prevent this issue.
Regards,
falcon