Configuring More than 4 script triggers to a NI AWG 5451

Solved!
Go to solution

Is it possible to configure more than 4 script triggers to a NI AWG 5451? The 5451 specifications document manual indicates that there are many possible trigger sources (e.g. two front panel connectors and 8 back-plane connectors) but the online help seems to indicate that it's only possible to configure 4 script triggers:

 

http://zone.ni.com/reference/en-XX/help/370524P-01/fgencref/cvinifgen_configuresoftwareedgescripttri...

 

Any help appreciated 😞

0 Kudos
Message 1 of 2
(2,288 Views)
Solution
Accepted by topic author aquirdturtle

Based on our documentation of script triggers it seems like the maximum number of script triggers is four. 

 

Script Triggers

http://zone.ni.com/reference/en-XX/help/371025P-01/rfsg/script_triggers/

 

If you search for the word "four" in this document you will find my reference. 

 

Spencer R | NI