03-03-2014 01:17 PM
Hey, I'm unsing 2 NI MIMO platforms, one with 4 VSG (NI 5673 E) modules (one master and 3 slaves) and the other with 4 VSA (NI 5663 E) modules (1 Master and 3 slaves also). By connecting the VSG Master to any VSA (Master or Slave), I've got a signal (after configurating them with MAX), BUT by connecting an VSG slave (5673 E) to any VSA, there 's no signal received and shown (on soft front panel) .
It seems to be a clock, problem ! If yes, how can I solve this ! If not, what is wrong ?
Solved! Go to Solution.
03-03-2014 03:05 PM
When you are using the 5673 with an external LO:
You must make sure that external LO signal is present. The master will not export the LO signal unless you tell it to.
Open two RFSG soft front panel windows, one to the master and another to the first slave in the LO daisy chain. One the master, click the 'Device/System' button, 'LO', 'LO Export' to export the LO signal.
You will need to do this on each 5673 between the master and the slave you are trying to use.
03-03-2014 03:18 PM
Thanks mbb25
I've just enabled the LO as suggested on RFSG front panel, but nothing changed when I just click on " RF On/Off" button of the slave RFSG front panel !
Should I change something else about the clk settings ?
03-03-2014 03:43 PM
Thanks mbb25, I made it with master and slave 2, and after trying with master and slave 1, it was ok.
Thanks for your help 😉