The device could not be calibrated -201295

Hello,

 

In one PXIe slot I have the DAQ card PXIe-6363 connected. When I press the button "Self-Calibrate" in MAX I get the next error message:

 

The device could not be calibrated

Status Code: -201295

Device could not complete the calibration operation

 

I have done that with load connected to the card and also with no cable connected to the card. Does anyone know how to solve that problem?

Much thanks for your time and dedication!
0 Kudos
Message 1 of 2
(3,238 Views)

Hi,

 

in the self test calibration the board measures a signal and compares it with an internal reference. The issue might be related to a malfunction in the measurement or in the internal reference. In any case, is very likely that the board needs to be repaired.

 

My suggestion is to test it with the DAQ Diagnostic Utility and contact NI support attaching the report of the Utility. 

 

Regards,

 

 