PXI-4145 current limit

Hi all ,

 

I am using SMU 4145. I have some register writes should be done with 100mA current limit and range but as I complete that I have to set to 100uA  current limit and range to make a measurement.

so I have to repeat this for 10 times in the loop.

 

problem here is if i configure 100uA after 100mA SMU will give an error saying minimum is 1mA and cannot set 100uA limit and range .. every time I need to step down to 1mA limit and 1mA range than need to set to 100uA ..

than again while register writes I have to change range to 100mA than limit to 100mA

 

This I cannot repeat for 10 times .

 

Are there any way to switch between the limits and range?

 

 

Best Regards
Manasa M
0 Kudos