From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

PXI

cancel
Showing results for 
Search instead for 
Did you mean: 

NI-PXI 5412 Script Mode

Solved!
Go to solution

Hi,

I am using the niFGEN Configure Output Mode VI to allow for two 5412 FGENs to produce waveforms with a specified phase difference.  However, I am finding that whenever I run the Configure Output Mode VI, I get an error if I choose "Script" as the output mode.  The error has error code -1074135039 and reads:

"niFgen Configure Output Mode.vi<ERR>Instrument error detected. Call the PREFIX_error_query function and examine the error code and error message parameters for additional information regarding the error."

Is the 5412 capable of script mode?  If so, what might I be doing wrong. (If so, I will be happy to post more info).

 

Thanks,

Billy Maier

0 Kudos
Message 1 of 2
(2,919 Views)
Solution
Accepted by topic author William.Maier

Look here: http://www.ni.com/pdf/products/us/2004_4600_305_101_DLR.pdf

 

Under "output modes," only " Arbitrary waveform mode and arbitrary sequence mode" are listed.  So the 5412 is not capable of script mode.  Compare to http://www.ni.com/datasheet/pdf/en/ds-235, for example.

Message 2 of 2
(2,906 Views)