Hi everyone, I am using the NI PXI 4065 DMM for a voltage measurement.  I am using the NI DMM READ.vi to read a single measurement.  I am setting the NI DMM resolution to 6 1/2 digits when I take my measurement.
I am trying to figure out if any averaging is done in a single measurement.  Per the NI 4065 spec (attached to this post), for a measurement with 6 1/2 digit resolution the aperature time is 100 NPLC.  Here are my questions:
1. What is the total amount of time it takes to take one measurement with the DMM set up to 6 1/2 digit resolution?
2. What is the default aperature time for 6 1/2 digit resolution?
3. How many samples are being averaged when the DMM is setup to 6 1/2 digit resolution with aperature = 100 NPLC?
Thanks so much!
(4,503 Views)
Yes, in a single measurement there is averaging done.
The amount of time it takes to take one measurement depends ont he aperature time. So, for example, if your power input to the chassis is a 50Hz line and you choose an aperature of 100 NPLC the time it would take is 1/50 *100 so this would be 2 seconds.
This document provides information on the default values for the 4065:
http://zone.ni.com/reference/en-XX/help/370384N-01/dmm/4065_dmm_defaults/
As you can see the default aperature time for 6 1/2 digits is 200ms
The number of samples per average is unfortunately not published.