06-26-2014 04:13 PM
Solved! Go to Solution.
06-27-2014 01:55 PM
I'm a little confused by your question. The manual indicates that you can connect to a load that has a floating ground, not that the current source can have a floating ground. All of the AO GND pins are indeed connected to the computer ground. To connect to a floating load, you would simply need to set up your circuit as in figure 2-4 from the manual. There is nothing to configure in software. If the load was grounded, I believe you would need to connect the load's ground to the AO GND pin.
06-27-2014 08:52 PM
The answer is no.
I spent the better part of the day on the phone with NI engineers. The PXI-6704 manual provides conflicting information. Indeed it informs the PXI-6704 operator that she "can connect to a load that has a floating ground," but it should also mention that the load will no longer have a floating ground, because the device ground planes are connected to the computer system's ground signal, which is connected to earth ground. I'm not sure why the verbiage about floating ground is included because there is no way to isolate the load with this device.
Our team returned the PXI devices to NI. The Application Engineer we worked with has spoken to R&D about clarification in the manual.