ivi step 34410

 I have been using  the ivi step for a DMM step in Teststand. When using a NI dmm to read a voltage the ivi steps works ok, but when using a Agilent 34410A DMM it does not give the correct reading back. I have downloaded the 34410a driver from NI and used this rather than the one from Agilent.

Is it the case that the NI_ivi_DMM step only works for a NI dmm and not any other manufacturer?

 

 

0 Kudos
Message 1 of 3
(4,114 Views)
No, the ivi step should work for any instrument with an ivi driver. How different are the readings? Is the configuration correct? You should be able to set a breakpoint before the read and check the setup on the 34410 front panel.
0 Kudos
Message 2 of 3
(4,111 Views)

Found the problem, the driver session was set to 'Simulate', which resulted in the DMM returning random numbers.