09-06-2006 01:00 PM
09-07-2006 08:38 AM - edited 09-07-2006 08:38 AM
Hi Heather,
There are 2 ways I can suggest, as follows:
Method 1.
If you have access to Labwindows/CVI, load modelsupport2.prj from:
C:\Program Files\National Instruments\TestStand 3.5\Components\User\Models\TestStandModels
This project file contains: modelpanels.uir , which contains the Serial Number dialogue box (panel)
displayed by the Teststand PreUUT callback.
If you then edit the Ring Control (PREVUUTLIST), at the right side of the serial number string control,
so that it is Initially hidden, i.e. select the appropriate check box.
Save the UIR file, then run Teststand.
The Serial Number dialogue box displayed by Teststand should no longer display the Ring Control
at run time, i.e. the Operator cannot use the history to fill out the serial number.
(I tried this using Teststand 3.5 and Labwindows 8.0.)
Method 2.
Create your own PreUUT model callback within your client sequence file, this will replace the standard
PreUUT callback found in the Process Model.
This method takes longer, but allows you the opportunity not only to resolve your current problem but
also capture other info from the Operator by adding further controls to the custom dialogue box you
create.
regards,
Gary.
Message Edited by Gary H. on 09-07-2006 04:43 PM
09-07-2006 12:06 PM
Hello Gary,
Thank you very much for taking the time to try out the solution. I have implemented the first solution you suggested, using LabWindows and it was fairly simple to do; your instructions were very clear and concise and I sincerely appreciate this effective suggestion. This has solved the issue with retaining serial number history.
Regards,