samboos,
There are two steps to getting this working. First you need to get a working route in Switch Executive in Measurement and Automation Explorer (MAX). You can use the test panel there to make sure that the route is working correctly.
Then in TestStand, you can set up the switching tab to connect the route you defined. Once that route is connected, you can define the life of that route. The most useful settings are Manual, where the route won't disconnect until you run a disconnect operation, and Step, where the route will automatically disconnect after the step has finished running.
Let me know if this helps!
Josh W. | National Instruments | Applications Engineer