10-29-2014 10:45 AM
I have a PCI 6229 M Series Daq board and a BNC 2110 Connector panel. The PCI 6229 spec says that it has 4 Analog outputs. I can access AO 1 and AO 2 just find. I need to access A0 3 or A0 4. Lab view doesn't error out but I dont get any signal out of AO3 or AO 4. Can I access all 4 Analog outputs on the BNC 2110 connector panel??
Solved! Go to Solution.
10-29-2014 11:06 AM
10-29-2014 12:30 PM
Kudos for your quick response. Could you please provide a link to the document that you are referring to. Everything I have been looking at just mentions that the board has 4 analog outputs. I will look for another way to accomplish my task.
10-29-2014 12:49 PM
10-29-2014 02:03 PM
Thanks again for your response. For those following this forum I have copied this information out of the installation manual and have reformated the information. Hope this helps others out.
Connecting Analog Output Signals
Use the BNC-2110 BNC connectors on the front panel to connect AO <0..7> signals to your DAQ
device. The number of connectors you use depends on your DAQ device and application.
E/M/S Series DAQ devices can only use the AO <0..1> BNCs.
AO Series DAQ devices can use the AO <0..1> and the AO <2..7> BNCs.
Refer to your DAQ device documentation for information about the use of these signals.