01-19-2010 04:36 PM
Hello Guys,
My question is provided in the topic, do you have any idea regarding taht? your help is appreciated.
Ela
Solved! Go to Solution.
01-19-2010 07:21 PM - edited 01-19-2010 07:28 PM
Hi Ela,
For almost all of the devices supported by DAQmx, you cannot. When all of the AO channels have the same ground, connecting them in series would short the channel outputs to ground, which is bad.
However, there is an exception: the SCXI-1124 module has channel-to-channel isolation, which allows the channels to be cascaded to output higher voltages: Can I Cascade the Voltage Outputs of an SCXI-1124?
With channel-to-earth isolated devices (such as the NI 926x or NI 623x), you may be able to cascade multiple devices together, but not multiple channels on the same device.
Brad
01-06-2015 08:25 PM
The NI 9269 has channel-to-channel isolation, so we can cascade multiple channels on the same device: http://www.ni.com/pdf/manuals/375099b.pdf
01-07-2015 01:51 PM
Hi JKSH,
Thanks for pointing out that the SCXI-1124 is no longer the only DAQ AO module that supports cascading multiple channels. In addition to the NI 9269 that you mentioned, which supports cascading 4 channels to output +/-40 V, the NI PXIe-4322 supports cascading 8 channels to output +/-128 V.
Brad