I try to use the RS232 RJ45 port on cDAQ-9137 chassis to comunicate with CSZ thermal chamber.
I use a RJ45 to DB9 converter (Digikey Part # 602-1506-ND) with a null-modem (crossover) cable (Digikey Part # TL642-ND) to connect between cDAQ and CSZ thermal chamber.
The CSZ thermal chamber use Modbus and only pin 2, 3, and 5 is present. I made the labview program to talk to the device and tested using my laptop with USB-COM converter. and the program works with no error.
Once I switch to cDAQ and use the same program. The error in MODBUS read presents. I shut down both cDaq and thermal chamber and restart them but it still won't work.
I used COM1 in cDAQ for this communication testing. The cDAQ also used a Virtual COM3 (VCP) to communicate with other device.
Seems like the port on cDAQ has some problem or I didn't configured it properly.
Any recommendation and advice is appreciated. Let me know if further information is required.
10-10-2016 12:51 PM
Hi Luyi,
Have you made sure to have installed NI VISA in your cDAQ?
Regards,
Daniel C.
10-10-2016 07:10 PM
Hi Daniel,
Thank you for the reply. in NI MAX, the COM1 is visible.
Eric
10-11-2016 10:44 AM
Hi Eric,
Try making a loopback test on the cDAQ's serial port to make sure it is configured properly.
http://www.ni.com/tutorial/3450/en/
Regards,
Daniel C.
10-11-2016 11:29 AM
Thank you for reply.
I compared the pinout from cDaq manual with the converter cable from Digikey. They are not the same.
cDaq uses 8 and 9 for TxD and RxD. However, EIA-232D dictates 5 and 6 for TxD and RxD.
I'm ordering converter cable from NI (#182845-01) to fix this. This is not the first time that a simple regular hardware off the shelf somewhere won't work on NI.
I will get the cable in a week and will try it out. Thank you all for the help.