08-25-2015 02:45 AM
Hi,
Im trying to test the USB-6210 for multi channel acquisition using DAQmx but ther is something strange about the time the device takes to collect data for single channel acuisition to multi channel acquisition: it comes out to be same!
For example in the following code I add single channel with sample per channel = 25000 and sample clock rate = 250000.
void startDataRead() {
noOfSamplesToCapture = 250000;
// Create a new task myTask = new NationalInstruments.DAQmx.Task(); // Create a virtual channels myTask.AIChannels.CreateVoltageChannel("Dev1/ai0", "Ch1Sensor", AITerminalConfiguration.Rse, Convert.ToDouble(-10), Convert.ToDouble(10), AIVoltageUnits.Volts); // Configure the timing parameters myTask.Timing.ConfigureSampleClock("", Convert.ToDouble(250000), SampleClockActiveEdge.Rising, SampleQuantityMode.ContinuousSamples, noOfSamplesToCapture); // Verify the Task myTask.Control(TaskAction.Verify); runningTask = myTask; analogInReader = new AnalogMultiChannelReader(myTask.Stream); analogCallback = new AsyncCallback(AnalogInCallback); // Use SynchronizeCallbacks to specify that the object // marshals callbacks across threads appropriately. analogInReader.SynchronizeCallbacks = true; analogInReader.BeginReadWaveform(Convert.ToInt32(noOfSamplesToCapture), analogCallback, myTask); testTimer.Start(); } private void AnalogInCallback(IAsyncResult ar) { testTimer.Stop(); try { if (runningTask != null && runningTask == ar.AsyncState) { data = analogInReader.EndReadWaveform(ar); System.Diagnostics.Debug.WriteLine(testTimer.ElapsedMilliseconds); testTimer.Restart(); analogInReader.BeginMemoryOptimizedReadWaveform(noOfSamplesToCapture, analogCallback, myTask, data); } } catch (Exception) { throw; } }
I measure the time between callbacks and it comes out to be almost 1 second. Okay thats logical, it takes 1 second to collect 250000 samples for 1 channel at 250000kHz rate.
But when I add 5 channels as follows:
void startDataRead() { noOfSamplesToCapture = 250000; // Create a new task myTask = new NationalInstruments.DAQmx.Task(); // Create a virtual channels myTask.AIChannels.CreateVoltageChannel("Dev1/ai0", "Ch1Sensor", AITerminalConfiguration.Rse, Convert.ToDouble(-10), Convert.ToDouble(10), AIVoltageUnits.Volts); myTask.AIChannels.CreateVoltageChannel("Dev1/ai1", "Ch2Sensor", AITerminalConfiguration.Rse, Convert.ToDouble(-10), Convert.ToDouble(10), AIVoltageUnits.Volts); myTask.AIChannels.CreateVoltageChannel("Dev1/ai2", "Ch3Sensor", AITerminalConfiguration.Rse, Convert.ToDouble(-10), Convert.ToDouble(10), AIVoltageUnits.Volts); myTask.AIChannels.CreateVoltageChannel("Dev1/ai3", "Ch4Sensor", AITerminalConfiguration.Rse, Convert.ToDouble(-10), Convert.ToDouble(10), AIVoltageUnits.Volts); myTask.AIChannels.CreateVoltageChannel("Dev1/ai4", "Ch5Sensor", AITerminalConfiguration.Rse, Convert.ToDouble(-10), Convert.ToDouble(10), AIVoltageUnits.Volts); // Configure the timing parameters myTask.Timing.ConfigureSampleClock("", Convert.ToDouble(250000), SampleClockActiveEdge.Rising, SampleQuantityMode.ContinuousSamples, noOfSamplesToCapture); // Verify the Task myTask.Control(TaskAction.Verify); runningTask = myTask; analogInReader = new AnalogMultiChannelReader(myTask.Stream); analogCallback = new AsyncCallback(AnalogInCallback); // Use SynchronizeCallbacks to specify that the object // marshals callbacks across threads appropriately. analogInReader.SynchronizeCallbacks = true; analogInReader.BeginReadWaveform(Convert.ToInt32(noOfSamplesToCapture), analogCallback, myTask); testTimer.Start(); } private void AnalogInCallback(IAsyncResult ar) { testTimer.Stop(); try { if (runningTask != null && runningTask == ar.AsyncState) { data = analogInReader.EndReadWaveform(ar); System.Diagnostics.Debug.WriteLine(testTimer.ElapsedMilliseconds); testTimer.Restart(); analogInReader.BeginMemoryOptimizedReadWaveform(noOfSamplesToCapture, analogCallback, myTask, data); } } catch (Exception) { throw; } }
Still the time between each callback comes out to be 1 second. And as I check the 'data' variable in each callback. it contains valid data of 250000 samples for each channel everytime.
I expected it to take 5 seconds between each callback in order to collect 250000 samples of each channel. Please help me out here. What am I doing wrong?
Thanks.
Regards,
Zeeshan
09-01-2015 03:35 PM
Everything is correct. 1 clock - all channels. Rate is specified in samples/s, not samples/ch/s
Ideally the board should collect samples on all channels simultaneously on every clock tick. If there are N ADCs on the board (simultaneous sampling type), each ADC samples at the same time.
USB-6210 has one ADC with multiplexer - it switches much faster than acquisition clock.
Some ADCs have hold curcuit to keep voltage until it is processed by ADC.