09-30-2016 07:31 AM
The below table is from the data sheet of NI-6343 module.
What exactly is this random noise? And why does it vary with voltage range?
12-09-2016 09:29 AM - edited 12-09-2016 09:33 AM
See this.
If a device has multiple gains, it multiplies an input signal by one of the gains to make the signal take up more of the full device range.
And this (press Ctrl+f and search for Random).
Random/System Noise—Additional system noise generated by the analog front end, measured by grounding the input channel
Essentially, each input range is related to a different gain. The Random Noise is affected by the gain. The larger the input range, the larger the gain, thus, the larger the Random Noise.
Hope it helps,
12-13-2016 12:52 AM
