Virtual Ground on NI 9217

I'm using an NI 9217 connected to a DAQ to take 4 wire measurements of resistance across a custom built platinum RTD. The RTD is exposed to a salty water solution a few micrometers away from a platinum wire that is also exposed to the salty water. The potentials on the platinum wire as they go through the water near the RTD are +6V to +5.75V from earth. I would like to verify that I can set "com" on the NI 9217 to +5V from earth to prevent current flow through the water. The spec sheet seems to indicate this works, but before I go and mess up the module I wanted to verify.

 

9217 question.png

0 Kudos
Message 1 of 5
(2,907 Views)

Hey estbiostudent,

 

Could you point me to the spec sheet that shows that this works

0 Kudos
Message 2 of 5
(2,801 Views)

NI 9217 - Data Sheet - National Instruments: http://www.ni.com/datasheet/pdf/en/ds-193

Page 6, section "Safety" 

 

Safety
If you need to clean the module, wipe it with a dry towel.
Maximum Voltage 
Connect only voltages that are within the following limits.
All terminals-to-COM ±30 V

0 Kudos
Message 3 of 5
(2,797 Views)

and then also "Available channel-to-earth ground double-isolation barrier for safety, noise immunity, and high common-mode voltage range" under "key features" found on page 2

0 Kudos
Message 4 of 5
(2,796 Views)

Yes, you are correct you will be able to set the "COM" to be +5V above earth! 